TRACERCO Radiation Monitor. Instruction and Maintenance Manual for Model No. T202

Similar documents
TRACERCO Radiation Monitor

TRACERCO T402 and T402HR USER MANUAL

Our Extended Personal Electronic Dosimeter (PED) Family

RAM GENE-1 ERK. Contamination & Radiation Meter. Operating Manual. Document #PRIR90N2.DOC Version 2.2 October 2009

For further details please contact:

Using the GammaRAE II Responder Personal Radiation Detector and Dosimeter in One. Firmware Version 3.00 D August 2006

Contamination Meter Users Manual

RAM GENE / RAM GENE-1

LUDLUM MODEL 25 AND 25-1 PERSONAL RADIATION MONITOR USER S MANUAL. October 2014 Serial Number and Succeeding Serial Numbers

User s Manual. Electronic Personal Dosimeter (For Gamma(X)-ray) NRF30. Fuji Electric Co., Ltd. TN514629i 1/31

ANALOX 101 D2 PORTABLE OXYGEN MONITOR USER MANUAL

HIn3550 Magnetic Field Monitor User's Manual

Using the GammaRAE II Personal Radiation Detector

Digital Electronic Thermostat With RF

Learning. Radiation Protection Compliance Safety Assessment

II 2 G. SENSIT HXG-2d. Ex ib IIB T3 IP20. Suitable for combustible gases such as Methane, Butane, Propane and Natural Gas MADE IN USA

EVac II Programmable Digital Vacuum Gauge

Digital Electronic Thermostat With RF

Personal Radiation Alarm Dosimeter LK Operating Instruction Manual. Contents

GasAlertMicroClip X3 Specifications

MODEL 23 AND 23-1 ELECTRONIC PERSONAL DOSIMETER. August 2017

ANALOX 5001 Carbon Dioxide Monitor. User Manual ANALOX Analox 5001 Carbon Dioxide Monitor User Manual

PORTABLE ISOTOPE IDENTIFIER Search Tool / Sample Counting System

SENSIT HXG-2d Combustible Gas Detector INSTRUCTION MANUAL Read and understand Instructions before use. PENDING Ex. II3G Ex ic llb T3 IP20

SENSIT HXG-2d. SENSIT TECHNOLOGIES 851 Transport Drive Valparaiso, IN Phone: (219) Fax: (219)

Dosimeters/Survey meters

MODEL 951 RADIATION DETECTOR

Dual-Diode Dosimeter Patient Dose Monitor

Real-Time Remote Monitoring System Utilizing New Electronic Personal Dosimeter

Radioactivity dosimeter RD01. User's Manual

6100 SINGLE LOOP DIGITAL ADDRESSABLE FIRE ALARM CONTROL PANEL

Analox 1000 Series. User Manual. Analox Sensor Technology Ltd. 15 Ellerbeck Court, Stokesley Business Park North Yorkshire, TS9 5PT, UK

LUDLUM MODEL 26-1 FRISKER USER S MANUAL. April 2017 Serial Number PF and Succeeding Serial Numbers

CASS9000 Standalone FingerPrint Door Access System. User Manual

Single Gas Clip and SGC Plus User s Manual Rev 2.1 READ FIRST BEFORE OPERATION WARNING STATEMENTS / AVERTISSEMENT Activating tthe detector for the

ANALOX 3000 Carbon Monoxide Monitor

ToxiRAE II. PGM-1100 Series Personal Toxic Gas Monitor. User Manual

Druck DPI 705/DPI 705 IS Series

MICROpeL 75 Combustible Gas Sensor Part Number: PM

DMC 3000 Quick User s Guide

Single Gas Clip and SGC Plus User s Manual Rev 2.3

IMR IX176 Portable Gas Detector User Manual

TS400. Operating Manual. Test Station for Microtector II Series (G450/G460)

KELCO F29 PROGRAMMABLE FLOW SWITCH

actoolsupply.com User Guide Extech RD300 Air Conditioning AC Refrigerant Leak Detector Refrigerant Leakage Detector Model RD300 actoolsupply.

4) Intrinsic Safety Certification

ID2000. operating manual

ToxiRAE II. PGM-1100 Series Personal Toxic Gas Monitor. User Manual

Dräger Pac Ex 2. Elegant and smart... Explosive hazards and lack or surplus of oxygen under control

Controls Surface mounted pushbuttons Suitable for gloved use. Buttons Side - On/Off Front Menu Scroll Pin Hole Setup access

RAD-60 USER GUIDE (ALSO FOR THE RAD-50 MODEL AND ISOTRAK DOSIMETERS)

Operating Guide. English Version 3.3

INSTRUCTION MANUAL (ATEX)

PERSONAL COMBINED RADIATION DETECTOR/DOSIMETER РМ1703МО-1 OPERATION MANUAL

PM1703MO-1A PM1703MO-1B PERSONAL COMBINED RADIATION DETECTOR/DOSIMETER. Operation Manual

ZX1e ZX2e ZX5e. Document No Issue 01 user manual

Instruction Manual. Self-Leveling Combination Cross-Line Laser and Five-Beam Laser Dot Model No , &

Safe Operation Manual. Honeywell IQ Force Gas Detector

Macurco Single-Gas XL Series Monitor, CM-1XL Carbon Monoxide (CO), HS-1XL Hydrogen Sulfide (H2S) User Instructions

Dräger X-am approved as type MQG 0010 Multi-Gas Monitor Technical Manual _01.eps

GAS-Mate. Combustible Gas Leak Detector

1.1 Definitions of alert icons. For your safety

INSTRUCTION MANUAL IS-mB1 Minialite Intrinsically Safe Round LED Beacon

Digital Room Thermostat

KANE-ALERT-CO 2 Monitor with Temperature

3500 CONVENTIONAL FIRE ALARM CONTROL PANEL

Instruction Manual. Self-Leveling Combination Cross-Line Laser and Five-Beam Laser Dot Model No &

Instruction manual EN

SF 6 -Leak Pointer Operating Manual (original version) Type-No.: R002

ALTAIR 5X Multigas Detector Product Specification

User Manual. Humidity-Temperature Chart Recorder. Model RH520

Nuclear Associates

SENSIT HXG-2d Combustible Gas Detector

NexSysLink. 2 CAN Display Operation Manual. CAN Instruments Product Family

Intelligent Security & Fire Ltd

INSTRUCTION MANUAL T119 1MN0110 REV. 0. operates with ISO9001 certified quality system

Model R9400. Instruction Manual. CO Meter with Temperature. wwwreedinstruments. com

275HVD NON-CONTACT, HIGH VOLTAGE DETECTOR DETECTOR DE ALTA TENSIÓN SIN CONTACTO

MicrOTDR. Optical Fault Locator. User Manual

GG-EM ENTRANCE MONITOR. Installation and Operation Manual

REFRIGERANT LEAK DETECTOR

OWNER S GUIDE TYPE A TYPE C TYPE B TYPE E TYPE F TYPE D COOL FAN OFF. CA Series << FAN SPEED COOL FAN AUTO SLEEP TIMER FAN SPEED

KFP-CF Series Operation Manual

Dräger X-am approved as type MQG 01** Multi-Gas Monitor. Technical Handbook. CO2 Vol% CH4 %UEG. CO ppm. ppm NH _01.

Model: RT310. Installation Manual

Instruction Manual. Self-Leveling Rotary Laser Level Model No &

User Guide. for the Beacon ProActTM 200 System

OPERATING AND MAINTENANCE MANUAL. for the G30 CARBON DIOXIDE MONITOR

Model: RT310RF. Installation Manual

(Place Logo from Lab here)

MODEL: SP12C7. (Portable 4 gas detector) Operating Manual

Digital Control Unit- Installation, Operation & Maintenance IMPORTANT INFORMATION

LUDLUM MODEL 26 FRISKER USER S MANUAL. December 2016 Serial Number PF and Succeeding Serial Numbers

IRceL CH4 Methane (CH 4

Ranger EXP RADIATION A L E R T

Laboratory Airflow Monitors & Controls TYPE C STANDARD & NIGHT SETBACK LED & DIGITAL - VAV SYSTEMS

Logbook. Precept EN Fire Detection/Alarm Panel. & Precept EN Repeater

Bacharach Bodyguard 4 User's Guide

2) This manual covers Fire and General Alarm systems. The differences are described in the appropriate sections.

DMC 3000 User s Guide

Transcription:

TRACERCO Radiation Monitor Instruction and Maintenance Manual for Model No. T202

Page 2 CONTENTS 1. General Description 4 2. Operational Safety 4 3. Post-Delivery Checks 4 4. Operational Features 5 4.1 Switch On 5 4.2 Speaker 5 4.3 Backlight 5 4.4 Low Battery 5 4.5 Pre-Set Alarm 5 4.6 Calibration 6 5. Measurement of Radiation Doserate 6 5.1 Procedure 6 5.2 Analogue Display 7 5.3 Bargraph Display 7 5.4 Digital Display 7 5.5 Accumulated Radiation Dose 7 5.6 Peak Radiation Dose Rate 7 5.7 Setting Alarm Trip Levels 8 6. Maintenance & Calibration 10 6.1 Battery Replacement 10 6.2 General Maintenance 10 6.3 Repair & Calibration 11 7. Information Database 11 8. Technical Specification 12 8.1 Additional Information on Monitor Calibration 13 9. Hints for Doserate Monitoring 13 10. Warranty 13 11. Health & Safety 14

DOSE RATE RADIATION MONITOR Page 3

Page 4 1. GENERAL DESCRIPTION The T202 is an Intrinsically Safe and battery operated radiation monitor, capable of measuring X and gamma radiation over a wide energy range down to 60keV. The monitor reflects Tracerco s 40-years of experience in Industrial and Environmental Radiation Measurement by combining novel design features with major advances in processor technology, to enhance versatility, accuracy and ease of operation in the field. The radiation detection capability is based on robust Geiger Muller tube technology located in the head of the monitor. The monitor has been specially designed to be lightweight but robust and comfortable when used over extended periods in the field. The outer protective case is manufactured from a high quality polymeric material to be chemically inert, sturdy and resistant to impact. The instrument has a particularly high tolerance to humidity and generally wet conditions. The monitor is a true all weather instrument and is suitable for use in demanding climatic and environmental conditions such as those found in the Mining and Oil and Gas Industries. 2. OPERATIONAL SAFETY The T202 Radiation Contamination Monitor is Intrinsically Safe to Ex ia IIC T4 and is suitable for use in Hazardous Area Zones 0, 1 and 2. Typical battery life is around 100-hours. A LO BATT warning appears on screen when the residual capacity reduces to about 4 hours. When the battery is exhausted, the screen clears. The screen will remain blank when the internal circuitry identifies that battery condition is below the minimum required for effective operation. PLEASE NOTE THAT BATTERIES SHOULD NEVER BE REPLACED WHEN WORKING IN A HAZARDOUS AREA. A selectable backlight facility allows measurements to be made under dark or low light conditions (note that the use of this function will significantly reduce battery life). An audible response is also available to complement visual measurements. In the event of a GM tube failure or circuit malfunction which interrupts the radiation detection signal, the screen will indicate the following warning message ERR1. 3. POST DELIVERY CHECKS All monitors supplied by Tracerco have been fully checked, tested and calibrated prior to shipping. The monitor will be fitted with a new battery and available for immediate operation. Each monitor is supplied in a protective overpack for transportation. All monitors should reach the customer in perfect working order. In the unlikely occurrence of a problem, please contact Tracerco immediately (contact numbers are provided within this manual).

Page 5 PLEASE ENSURE THAT THE MONITOR RADIATION TEST AND CALIBRATION CERTIFICATE IS PRESENT IN THE PACK. (Certificates should be filed in a safe place for future reference. In the event of difficulties, duplicate certificates can be provided by Tracerco free of charge). 4. OPERATIONAL FEATURES The operator may also wish to refer to the functionality image in the front of this manual. In certain parts of the world, natural background radiation values may be anomalously low e.g. Middle East and the operator may conclude that the monitor is non-operational. To minimise this risk, the monitor is programmed to recognise the following situations: 1. No recorded counts over a period of 20 seconds the monitor will display a lo BGnd message on the screen. 2. No recorded counts over a period of 40 seconds the monitor will recognise a detector problem and will display an ERR message on the screen. 4.1 Switch On The monitor is activated by pressing the red ON/OFF switch. This has a toggle action press for ON and press again for OFF. In the absence of any significant local radiation sources the analogue meter will register incident background radiation which will vary with geographical location. A typical background reading might be expected to be in the region 0 to 0.5 µsv/h (See note above). (Please note that there may be a slight delay between switching off the monitor and the screen clearing of information). 4.2 Speaker Pressing the GREY switch with the speaker symbol activates the internal speaker. The speaker provides an audible clicking sound whose rate is proportionate to the amount of radiation registered by the detector. The speaker may be silenced by re-pressing the GREY switch. The speaker symbol shows on the screen when the audible option has been selected. Battery life is extended with the speaker off. 4.3 Backlight The monitor display is fitted with a backlight which is available for use in situations of low ambient light intensity. The GREY switch marked with the sunburst symbol is pressed to activate the backlight facility. Since the backlight significantly increases the current drain on the battery, the pushbutton switch is designed for contact operation only ie., depressing the switch activates the backlight and releasing the switch cancels the light. 4.4 Low Battery The monitor is equipped to display a LO BATT message although it is recommended that the battery is replaced as soon as this message is activated, it might be anticipated that under normal conditions, the monitor is capable of operating for a further 4 hours. NB: THE BATTERY MUST NEVER BE REPLACED IN A HAZARDOUS AREA.

Page 6 4.5 Pre-Set Alarm The monitor is equipped with an alarm function which the operator can, within prescribed limits, pre-set at any required level of doserate or integrated dose. The alarm provides a continuous high frequency tone whenever the measured radiation doserate or integrated dose exceeds the value set by the operator. The alarm may be deactivated by the operator There is no mute (accept facility) for the alarm. Alarm activation is indicated by a flashing loudspeaker icon. To set the alarm trip level, the operator should refer to section 5.6. 4.6 Calibration Current UK legislation requires that all operational radiation monitors are subject to annual inspection and testing. This requirement includes performance checks and where appropriate, recalibration of the monitor. On switching on the radiation monitor, the next calibration due date will register automatically on the main screen for approximately 2 seconds. This date will indicate the 12-month anniversary of the current calibration. All Tracerco radiation doserate monitors are checked and calibrated immediately prior to despatch to the customer. Following this initial procedure (or the annual inspection and calibration check), Tracerco operatives will re-set the next calibration due date within the software of the monitor. Further information on the calibration procedure is provided elsewhere in this manual. 5 MEASUREMENT OF RADIATION DOSERATE (It may be a help to read this section in conjunction with the function flow chart) 5.1 Procedure Switching on the monitor will display all available segments for 2 seconds. A check should be performed for any missing segments. For the following two seconds, the next CAL DUE date will show using a MM/YY format. The monitor then enters normal operation. All previous information on peak radiation dose rate and accumulated radiation dose will have been re-set to zero following the power down.

Page 7 5.2 Analogue Display This reading is directly based on the analogue signal from the radiation detection circuitry and intended to provide an improved visual indication of doserate trends in the region 0 1000 µsv/h. The analogue display will remain at full scale deflection pending the removal of the radiation field. 5.3 Bargraph Display This display provides an improved visual indication of doserate trends in the region 0-1000µSv/h. It would be expected that in the majority of circumstances, the operator would record definitive doserate data based on the digital display. If the bargraphs full scale display is exceeded (1000µSv/h), the digital display will continue to operate up to a maximum of 10,000µSv/h. The bargraph display will remain at full scale deflection pending the removal of the radiation field. 5.4 Digital Display This display utilises a digitalised signal and is self-ranging, providing a continuous numeric display of doserate from 0 to 10,000 µsv/h. If the maximum radiation reading (10,000µSv/h) is exceeded on the digital scale, the display will show the following warning OVER. In this situation the monitor cannot provide a measurement of radiation dose exposure. If the radiation field exposure is reduced below 10,000µSv/h, the monitor will return to normal operation. 5.5 Accumulated Radiation Dose An important feature of the monitor is the capability to record accumulated dose. Once the monitor has been switched on, the accumulated dose will be continuously updated in the monitor memory. The yellow MODE button allows the operator to view the accumulated dose in µsv on the digital display. The button allows the operator to toggle between digital readout of doserate µsv/h and accumulated dose µsv. The operator can control the period over which the dose is integrated by simply pressing the SELECT / RESET button when in the accumulated (µsv) mode. The accumulated dose will return to zero and then automatically commence another period of dose integration. This accumulated dose function allows the monitor to be effectively used as an integrating dosimeter and is useful in carrying out assessments of potential dose uptake by individuals during the performance of certain operations. 5.6 Peak Radiation Dose Rate The monitor has the capability to record the highest doserate it has been exposed to since it was switched on. Switching off the monitor clears the memory of the previously held peak value. Pressing the green PEAK button while in the doserate (µsv/h) mode will show the peak dose rate. A second press will return the display to normal reading mode. Pressing the SELECT / RESET button during this peak display will reset the peak value to zero, allowing a new peak reading.

Page 8 5.7 Setting Alarm Trip Levels The T202 is equipped with an alarm function which can be set by the operator. The alarm is a continuous high-pitched note and operates whenever the monitor reading exceeds the pre-set trip level. The alarm trip level can be set in either dose equivalent rate values or as integrated dose equivalent. The incremental settings for the alarm are as follows:- a) Integrated Dose Equivalent Function Range µsv Increment µsv 0 10 1.0 10 400 10.0 Max alarm trip level : 400µSv. b) Dose Equivalent Rate Function Range µsv/h Increment µsv/h 0 50 1.0 50 100 5.0 Max alarm trip level : 100µSv/h. Sounder Options If the speaker symbol is pressed and released quickly, the speaker symbol will flash repeatedly at approximately 1-second intervals if the alarm setting is ON. The sounder should operate, providing audible clicks which are consistent with the level of radiation being measured. If the alarm setting is OFF, the speaker symbol will remain fixed on screen and the sounder will continue to operate as above. If the speaker symbol is absent from the screen, it indicates that both audible response and alarm setting is OFF. It should be noted that on switching on, the monitor will retain the previous alarm trip level settings and speaker status. To Set the Alarm Trip Level The following flowchart identifies the process for setting an alarm trip level in either µsv/h or µsv mode. (NB: Pressing peak increases the trip setting (+) and pressing mode decreases the trip setting (-)).

Page 9 T202 FUNCTIONAL DIAGRAM PRESS AND HOLD FOR 3 SECONDS ALARM SETTING FOR usv/h ALARM ON ( FLASHING) ALARM 0FF ( NOT FLASHING) DISPLAY usv/h DISPLAY MODE A-ON A-OFF RESET RESET INCREASE usv/h ALARM VALUE DECREASE usv/h ALARM VALUE PEAK MODE DISPLAY usv/h VALUE RESET usv/h ALARM ON usv/h ALARM OFF NOTE THAT THE ALARM LEVEL WILL ONLY OPERATE FOR THE DISPLAY MODE SELECTED. E.G. AN ALARM LEVEL SELECTED IN usv MODE WILL NOT SOUND IN usv/h MODE usv/h SWITCH ON ON / OFF ALL SEGMENTS LIT FOR 2 SECONDS IF RESET BUTTON PRESSED usv VALUE RESET. CAL DUE DATE SHOWN FOR 2 SECONDS RESET usv/h DISPLAY MODE usv DISPLAY PEAK PRESS AND HOLD FOR 3 SECONDS usv/h PEAK DISPLAY RESET ALARM ON ( FLASHING) ALARM 0FF ( NOT FLASHING) PRESS THE RESET BUTTON TO RESET THE PEAK VALUE DISPLAY usv DISPLAY MODE A-ON A-OFF usv RESET RESET INCREASE usv ALARM VALUE DECREASE usv ALARM VALUE PEAK MODE VALUE DISPLAY usv RESET usv ALARM ON usv ALARM OFF ALARM SETTING FOR usv SYMBOL OFF CONTINUOUS SOUNDER OFF ON ALARM OFF OFF SHORT FLASH OFF ON LONG FLASH ON ON

Page 10 6. MAINTENANCE & CALIBRATION The T202 is an Intrinsically Safe Radiation Monitor incorporating many design features which are based on Tracerco s 40-years of global experience in manufacturing radiation detection equipment for the chemical and both on-shore and off-shore oil and gas industries. The monitor has been designed for reliable operation in relatively hostile environments and should provide many years of effective service provided attention is paid to basic operational maintenance. 6.1 Battery Replacement The condition of the battery is constantly monitored. Normal battery life is approximately 100 hours for continuous or intermittent operation. With about 4-hours of battery life left, a LO BATT warning appears on screen. When the battery power is too low for reliable operation, the screen will blank. BATTERIES MUST ONLY BE CHANGED IN A SAFE AREA. The internal battery compartment is clearly identified in the rear of the case. The only batteries permitted are the alkaline manganese MN1604 or MX1604. These batteries are intrinsically safe for the gas groups and temperature for which the monitor is approved. Others are not. DO NOT USE ANY OTHER TYPE OF BATTERY. The battery cover can be removed by the single retaining screw. The gasket seal is attached to the inner face of the cover for additional safety. This seal must not be damaged during removal and must be kept clean at all times. If the seal is obviously damaged, further guidance should be sought from the Tracerco helpline (contact details are provided on page 14 of this manual) or via the Tracerco website under monitor sales, calibration and repair : www.tracerco.com. IMPORTANT: THE BATTERY COMPARTMENT MUST ONLY BE OPENED IN A SAFE AREA. 6.2 General Maintenance The liquid crystal display screen is sealed within the upper section of the case which provides a robust leak-tight fitting. The analogue meter has been specially selected for reliability in the field and the case is constructed from a lightweight impact and chemical resistant polymer. The monitor complies with ATEX requirements for use in hazardous area zones 0, 1 and 2. The only routine maintenance which is recommended is mild cleaning of the case and display to prevent the build-up of grease and other potential contaminants. DO NOT USE SOLVENTS ON THE WINDOW.

Page 11 6.3 Repair & Calibration The T202 instrument incorporates special software for checking and calibrating the response of the monitor to radiation fields. In order to comply with UK statutory regulations, the monitor must be inspected and tested on at least an annual basis. It is essential that such inspection and testing is carried out by Tracerco or an approved organisation working on their behalf which has access to the appropriate diagnostic software and has the necessary training and experience to ensure that the Intrinsic Safety features of the instrument are not compromised during the procedure. In addition to checking the physical condition of the monitor, Tracerco will check the response of the monitor to known gamma radiation fields of the isotope Caesium-137. The response of the monitor to low-energy gamma radiation fields will also be checked using a source of Americium-241. The checks on gamma radiation response will be carried out using a specially designed calibration rig incorporating two sealed sources of Caesium-137. This rig has been independently checked for absolute calibration (±5%) by a recognised competent authority and allows checks to be made over the full working range of the monitor. If appropriate, the monitor will be re-calibrated using the specially configured internal software. Battery replacement is a standard element of this annual inspection procedure. USA OPTION A modified version of the T202 monitor is available which has been specially configured for use in the USA. The instrument is direct reading in mrem units and must be specified when placing an order. This monitor retains all of the same functionality as the standard T202 version. 7. INFORMATION DATABASE To assist in providing customers with a more efficient and comprehensive service, Tracerco has developed a special database which holds records of all monitors supplied to customers, including details of inspections, repairs, replacements and calibrations. This database provides a life history for the tracking of radiation monitors.

Page 12 8. TECHNICAL SPECIFICATION Radiation Detected : X-rays and gamma rays in range 60keV to 1.25Mev Sensor : Single halogen, energy compensated Geiger Muller tube. Range Doserate (non USA) : Analogue display / Digital Bargraph 0 1000 µsv/h Digital display 0 10,000 µsv/h USA Version : Analogue display / Digital Bargraph 0 100 mrem/h Digital display 0 1,000 mrem/h Accumulated Dose (non : Digital display 0 10,000 µsv USA) USA Version : Digital display 0 1,000 mrem Peak Radiation Doserate : Digital display 0 10,000 µsv/h (non USA) USA Version : Digital display 0 1,000 mrem/h Over-range Response Analogues display / Digital Bargraph - will read hard full scale Digital display at over 10000µSv/h shows 0UEr (over) Variation with Temp : Less than ±10% over temperature range -10 C to +40 C Humidity Range : 0 95% Variation with Battery Voltage : Less than 2% Battery Life : 100-Hours typically with background radiation Low Battery Indication : On 4 hours available life left. Battery Alkaline Manganese MN1604 or MX1604. Case material Static dissipative nylon body with ABS window. The window is protected with a static dissipative additive Sealing Sealed to IP65 Certification : Intrinsically Safe Ex ia IIC T4 Ga (-20 C Ta + 40 C). Equipment category II 1 G Suitable for Hazardous Area zones 0, 1 and 2. Standard compliance The monitor meets the following EU directives 2004/108/EC Electromagnetic Compatibility Directive (including amendments) 94/9/EC ATEX Directive Weight : 600 grammes (approx.)

Page 13 8.1 Additional Information on Monitor Calibration Tests involving the response of the T202 monitor to radiation doserate have been carried out using a specially designed calibration rig which allows the instrument to be exposed to a range of radiation dose equivalent rates by precisely varying the distance from two independent Caesium-137 sources. The accuracy of the calibration facility has been independently checked and shown that the Ambient Dose Equivalent Rate deviation from 1/r 2 is less than 5% relative for the sources of Caesium-137 and Americium-241. 9. HINTS FOR RADIATION DOSERATE MONITORING Ensure audible output is selected when monitoring radiologically unknown environments. Minimises potential for exposure if attention is diverted from the monitor screen and provides an audible warning to other workers in the locality. Radiation detector is located parallel to the head of the monitor (approximately below the position of the Tracerco logo). For maximum accuracy and sensitivity point this face directly at the suspected source of radiation. Remember time and distance rules to minimise personal exposure. The accumulated dose and peak doserate options should reinforce safe working practices. Hold the monitor at a comfortable distance in front of the body. This approach maximises the response time to allow minimisation of the whole body dose. Wherever possible, check that the monitor is working before use by placing it in the vicinity of a known test source. Alternatively, ensure that sensible background radiation doserate values are being displayed. 10. WARRANTY It would be expected that under prescribed usage the T202 monitor should give years of reliable operation. If, however, any fault arises with the monitor within twelve months of purchase and Tracerco can reasonably attribute this to a manufacturing defect, Tracerco will repair or replace the monitor free of charge. If a fault occurs outside of the warranty period or if the monitor suffers damage as a result of an accident or inappropriate usage, Tracerco is able to provide a rapid repair service (usually within 7-days of receipt).

It is recommended that if the customer considers that there is a problem with the monitor, they should, in the first instance, contact our customer service helpline at the following address:- Tracerco Monitor Services Unit 2, Belasis Court Belasis Hall Business Park Billingham Cleveland TS23 4AZ UK Tel: +44 (0) 1642 375171 Fax: +44 (0) 1642 562819 E-Mail : adam.golightly@tracerco.com Page 14 Please note that Tracerco cannot accept liability for any loss or damage during shipping and the customer should ensure that the instrument is suitably packaged. 11. HEALTH & SAFETY The T202 Radiation Doserate Monitor has been tested and is certified for Intrinsic Safety, Coded Ex ia IIC T4 (-20 C Ta +40 C), Equipment category II 1G suitable for Hazardous Areas zones 0, 1 and 2. Tracerco considers that it has taken all reasonable precautions to ensure that when used as prescribed, the monitor will not endanger the health and safety of the operator. It is important, however, that the following basic guidance should be taken into account by the operator when carrying out any form of radiation doserate monitoring. 1. Operators should receive appropriate practical training in the use of radiation doserate monitors. 2. Battery life should be checked before use, particularly where work is to be carried out in a hazardous area. ON NO ACCOUNT MUST BATTERIES BE CHANGED IN SUCH AREAS. 3. The monitor has been designed and tested to EN60846:2004 to ensure that if the monitor is exposed to radiation levels well in excess of the full scale values (ie., up to 100 times), the readings will recover and normal operation resumed once the source of radiation has been removed. The digital scale will read OVER when maxima values have been exceeded and the analogue meter will sustain full scale deflection. The operator should never attempt to estimate a reading when over-range has occurred. 4. The monitor meets EN60846:2004 standard requirements for directional response. It is recommended that in normal use, the monitor is held with the scale horizontal and the audible response is operational. The operator should always be aware that exposure shine paths may exist where there is a marked variation in radiation doserates in both horizontal and vertical planes. It is strongly recommended that all operators receive adequate training for the type of radiation monitoring to be carried out.

Page 15 5. In the event of a Geiger Muller tube failure an ERR1 message will appear on the screen. In circumstances where the tube receives negligible background signal over a continuous counting period of 20 seconds, (and the tube remains operable), the monitor will display a lo BGnd message. If the monitor is exposed to a positive count rate the message will disappear and normal operation resumed. ESSENTIAL SAFETY INFORMATION HAZARDOUS AREA USE Marking details Certification number Protection Code Equipment category Baseefa05ATEX0151 Coded Ex ia IIC T4 Ga (-20 C Ta +40 C) II 1 G Hazardous area use The equipment is suitable for surface industry use in Zones 0,1, and 2 for gas groups IIA, IIB and IIC having a gas temperature classification of T1, T2, T3 and T4 in ambient temperatures of -20 C to +40 C inclusive. Practical safety points Do not open the battery compartment in a Hazardous Area. Use only the Duracell batteries MX1604 Alkaline manganese Or MN1604 Alkaline manganese Lock the battery compartment by the screw in the compartment lid. Do not over tighten. Do not use solvents on the monitor window.

TRACERCO and TRACERCO Diagnostics are trademarks of the Johnson Matthey Group of companies. Tracerco is a trade name of Tracerco Limited. 5th Floor, 25 Farringdon Street, London, EC4A 4AB. Company No.4496566 Tracerco Limited is a subsidiary of Johnson Matthey PLC. Tracerco Monitor Services Unit 2, Belasis Court Belasis Hall Business Park Billingham, Cleveland TS23 4AZ UK Tel: +44 (0) 1642 375171 Fax: +44 (0) 1642 562819 Tracerco 4106 New West Drive Pasadena, TX 77507 USA Tel: +1 281 291 7769 Fax: +1 281 291 7709 Toll Free: +1 800 288 8970 Tracerco Asia Sdn Bhd Lot 27, Jalan PJU 3/47, Sunway Damansara Technology Park, 47810 Petaling Jaya, Selangor, Malaysia Tel: +603 7803 4622 Fax: +603 7803 9312 TRACERCO and TRACERCO Diagnostics are trademarks of the Johnson Matthey Group of companies. www.tracerco.com tracerco tracerco.com XM0156/0/D